UV Signal just seems to be drifting

Posted: Wed Oct 18, 2017 9:05 pm
by sga
Hello all,

I am trying to requalify my 1100 with a UV DAD detector which has been lying idle for some time. The UV signal seems to be just mandering around betwen ±50 mAV and even though I have balanced the signal many times the baseline just does not seem to stabelise. Almost a sinusoidal wave.

Any idea what could be causing this ?

I have tried to inject Caffeen a few times to get a peak, but this also does not seem to be helping. No peak observed nothing.

Re: UV Signal just seems to be drifting

Posted: Thu Oct 19, 2017 5:56 am
by Gerhard Kratz
I guess your flow cell makes the problem, dirty, air bubbles, I would recommend first to rinse the flow cell and purge carefully with a syringe. Isopropanole to rinse is good. Good luck.

Re: UV Signal just seems to be drifting

Posted: Thu Oct 19, 2017 11:49 pm
by tom jupille
Do you see the same drift / wander with no flow?
And if it looks somewhat "sinusoidal" does the period change when you change the flow rate?
And does the problem vary with wavelength?

Re: UV Signal just seems to be drifting

Posted: Fri Oct 20, 2017 1:41 pm
by Steve Reimer
How many hours on the lamp?
How long has it been on?

Re: UV Signal just seems to be drifting

Posted: Thu Oct 26, 2017 11:49 am
by BostonFSE
observe the flow on the live plot then turn the flow off. If the baseline goes flat you have a dirty flow cell. If the baseline stays "wavy" you have a bad lamp
both are simple to replace but cost over $1000 each
Lamps are usually rate to 2000 hours or so