Peaksimple GC-FID upsidedown chromatogram
Posted: Wed Oct 18, 2017 2:47 am
by Jackkelly
My peaks are in the negative mV. I know the chromatogram is upside down completely because all the peaks are negative. I think its the peaksimple software. Please share any info you got on this.. Thanks for the assistance.
Jack
Re: Peaksimple GC-FID upsidedown chromatogram
Posted: Wed Oct 18, 2017 8:27 am
by dblux_
I don't know your specific hardware configuration, but you may check whether your detector allows to change the polarization of the output signal.
Re: Peaksimple GC-FID upsidedown chromatogram
Posted: Wed Oct 18, 2017 9:48 am
by tkubowicz
Hello
If you have cable from GC (signal output) to your PC (or signal converter) try to swap wires to change polarity - check manual for configuration.
Regards
Tomasz Kubowicz
Re: Peaksimple GC-FID upsidedown chromatogram
Posted: Wed Oct 18, 2017 8:57 pm
by Jackkelly
Resolved:
The polarity was swticed as suggested. I just switched the two wires on the data system. Thanks for the assistance.