Posted: Tue Oct 17, 2017 1:11 pm
by chloesabrina
Hi All, I just cleaned my EI source (first time for an extractor source) and clipped the column on the GC-MS interface. Since there is a cone attached at the interface how much should the column show past the transfer line? I can't find anywhere on the web or in the manual that addresses this. I figured asking her is normally faster than trying to get the information from Agilent.
Re: Agilent 5977 with extractor source
Posted: Tue Oct 17, 2017 1:29 pm
by James_Ball
About 1-2mm past the inner sleeve of the interface just as with a normal source. You can loosen the setscrew on the cone and slip it off, install the column as usual then replace the cone. I found that to be the easiest way to do it.
Re: Agilent 5977 with extractor source
Posted: Tue Oct 17, 2017 1:33 pm
by chloesabrina
Thanks James - see I knew it would be faster on here