Determination of quaternary ammonium preservatives
Posted: Sun Dec 03, 2006 10:07 am
by mross
I have a quaternary ammonium salt with a UV at 205nm. It is unretained on RP without ion-pairing. The problem is that with all the ion-pairs that I know they have strong UV adsoprtion.
ANy ideas,
Posted: Sun Dec 03, 2006 5:02 pm
by Uwe Neue
Depends on the concentration that you need to detect. There is a technique around with UV-absorbing ion-pair reagents, where you look for the concentration of the ion-pair against the ion-pair background to detect the analyte. You could use that trick, but I doubt that the sensitivity is very high.
Alternatively, you could try HILIC with a silica column and an acetonitrile-phosphate buffer mobile phase.
Posted: Mon Dec 04, 2006 3:54 am
by SIELC_Tech
Hi Malcolm,
Here are few applications for retention/separation of quaternary compounds on Primesep columns. No IP reagent required. Compounds are retained based on reverse phase and cation exchange (anion exclusion) mechanisms:
http://www.sielc.com/compound_080.html
Contact me directly if you need more information or would like us to develop method for you (free of charge).
Vlad
Posted: Mon Dec 04, 2006 5:49 pm
by Mark Tracy
You can try 10mM HClO4 as an ion pairing agent; it ion-pairs about as well as TFA, and has much better UV properties. The alkane sulfonic acids have good UV properties and are stronger ion-pairing agents. You can also look into evaporative light scattering or suppressed conductivity detection. I have done quaternary surfactants like benzalkonium chloride on an Acclaim Surfactant column with conductivity detection, and it gets better detection limits than ELSD and sometimes better than UV.
Posted: Thu Dec 07, 2006 3:18 am
by ym3142
mross,
the common used IP are alkane sulfonic acid and tetraalkylammonium. I do not have an impression of either one having strong UV.
Could you let me know what kinds of IP you tried?Thanks
Posted: Fri Dec 29, 2006 2:56 pm
by dpr
I realise it's a while since this was posted, however.
We have some experience of quartenary ammonium salt determinations by IC.
Iodine salt of Bis-Dibutyl Ethyl Ammonium Hexane is analysed on Dionex NS1 columns.
MSA eluant with methanol solvent.
Drop me an email if you want any information
Posted: Wed Jan 10, 2007 11:46 pm
by Wen Jiang
mross,
Another possibility is using HILIC. One typical example can be found in the following web address:
http://www.sequant.com/sn/ufiles/SeQuan ... 00-14A.pdf