lifetime of filaments
Posted: Sat Dec 02, 2006 8:56 pm
by Anica
I would like to know an average lifetime of filaments used in GC/MS and what factors it depends on (only the solvent cut time?). We have Shimadzu 2010 QP. First filament lasted only 255 h of 1500 h supposed. Any suggestions as to why the filament might have blown so suddenly would be appreciated. Anica
Posted: Sat Dec 02, 2006 10:16 pm
by nolteboy
Not being familiar with Shimadzu MSD's (only HP/Agilent), I'd hafta say that 1000 to 1500 hours is about right in my experience. If this is indeed your first filament, it might have just been a statistical fluke (i.e. that particular filament might have been a "lemon"). You have checked the filament voltage setting to make sure that it's not in excess of what would be reasonable (~70 electron volts or thereabouts is typical for most applications)? Your system is tight and leak-free? (besides contributing to high baselines and poor sensitivity, too much air in system could shorten filament lifetime).
Posted: Wed Dec 06, 2006 3:22 pm
by vperica
Very important factor for life time filament is level of water and air.Check
gas purifier for moisture and oxigene.
Perica
Posted: Tue Dec 19, 2006 12:29 am
by james little
Scientific Instrument Services rebuilds filaments for a good price and does a good job. Often better than the original from the manufacturer.