Please give all the details for the set-up - concentration of benzene and naphthalene in the methane, volume injected, column, flow rates etc etc.
Chrompack 9001
CP-Sil 5CB, 0.32 mm, 40 m
35 kPa H2 carrier
loop volume appr. 50 microL (gas)
0.5 % C6H6, 0.015 % C10H8, balance CH4
analysis: 8 min at 40°C, ramp 50°C/min tp 260°C, 10 min at 260°C
Does the ratio of methane peak area to benzene or naphthalene conform to the expected ratio ?
CH4 is measured in one range (less sensitive) because of its concentration, and the rest of hydrocarbons in a more sensitive range, but both C6H6 and C10H8 are measured in the SAME range.
The link to the chromatogram does not open, the shoulder might point to something eluting with the naphthalene and increasing its area, what happens if you run just the methane ?.
Please, download the picture.
Methane alone nor benzene/CH4 mixtures shows such a feature.
Benzene-to-naphthalene area ratio is hold aproximately constant at various dilutions in the methane stream.
Area ratios benzene-to-CH4 and naphthalene-to-CH4 seem to be linear with respect to molar ratios in the sample.
Another possibility is that the benzene peak is off the top of the detector or electronics linear range.
Neither C6H6 not C10H8 peaks reach the saturation level at that range.
I cannot imagine why this is happening !!!