USP <41> Minimum Net Weight
Posted: Fri Oct 13, 2017 7:38 pm
by dchau503
Hi all,
I have a question about USP <41> and the minimum weighing criteria. The balances usually say that the minimum net weight is x; therefore weights below x are not accurate and would be in violation of my SOPs. If y is greater than x, mm I able to get around the minimum weight by putting in a container with weight y+x and then subtract out y? y+x and y are both greater than x.
Re: USP <41> Minimum Net Weight
Posted: Mon Oct 16, 2017 4:46 pm
by DR
In a word, NO!
Let's say you have a 4 place balance. Even if your RSD for replicate 50mg weights is 0, the 0.41 rule limits your accurate weight limit to >85mg.
Weighing less than that in a 2g weigh boat makes no difference, your net weight has to be at least 85mg to possible be "accurate" per the USP.
Posted: Mon Oct 16, 2017 9:44 pm
by Peter Apps
Hi all,
Somewhere in the archives is a thread about this exact subject, and it goes on and on for about 20 pages. The answer is short and simple though; NO.
Posted: Tue Oct 17, 2017 10:53 am
by HPLC chemist
I agree with Peter! You cannot squirm your way out of compliance with the regulations. Buy an ultrabalance!