I have a L35 column and a RID detector, but the areas of the peak of zoledronic acid are so small.
Somebody have experience in these product...?
Thanks
Elsewhere you say you have an L23, which I think is an anion-exchange quat ammonium column. The sensisitivity will depend on the stability of your detector, mobile phase, baseline, and the amount injected. For RI, temperature stability is critical to obtain stable baselines. Some of these bisphosphonate compounds can be difficult to quantitatively elute off ion-exchange columns using isocratic systems.
If the method worked before, what has changed?. If the method is new, what sensitivity do you want, and how close to that are you?.
If you are well away from the desired sensitivity, and the Ri is working to spec, then I would recommend using another method.
There are UV and ELSD reverse phase methods using tetrabutylammonium ion-pairing, listed on Pubmed. ust search on HPLC and Zoledronic Acid.
eg
"Separation of zoledronic acid and its related substances by ion-pair reversed-phase high performance liquid chromatography"
Zhang X, Jiang Y, Xu Z.
"A rapid and simple ion-pair reversed-phase high performance liquid chromatographic method (HPLC) has been established for the routine analysis of zoledronic acid and its related substances. ... The excellent separation of zoledronic acid from its related substances, including the remaining imidazol-1-ylacetic acid used in the synthesis of zoledronic acid and other impurities of oxidation and decomposition, was achieved within 9 min on a Hypersil C8 column with UV detection at 220 nm. The mobile phase was a mixture of methanol (20%) and 5 mmo/L phosphate buffer (80%) that contains 6 mmol/L tetrabutylammonium bromide. The resolution factor of zoledronic acid from its adjacent peak was more than 2.5. This is a simple and rapid method for the routine assay of zoledronic acid. "
Bruce Hamilton