				Aerosil in sample
				Posted: Fri Nov 24, 2006 12:32 pm
				by Mattias
				Hi,
Does anyone have experience in removing Aerosil (water-free colloidal silicon dioxide) from samples? The solutions become cloudy, and the filters (0.22 µm) get stuck immediately. Even high-speed centrifugation does not help.
			 
			
					
				
				Posted: Fri Nov 24, 2006 5:05 pm
				by Uwe Neue
				I would acidify the sample to precipitate the colloidal silica. Watch out for loss of analytes.
			 
			
					
				
				Posted: Sat Nov 25, 2006 7:17 am
				by Russ
				Have you tried filters with a glass fiber prefilter?
			 
			
					
				
				Posted: Mon Nov 27, 2006 7:27 am
				by Mattias
				The solutions are cloudy even after flitration with 0.22 µm filters (the two drops that pass through the fliter)
I will try to acidfy the samples, haven't tried that yet. Thanks for the help!
			 
			
					
				
				Posted: Mon Nov 27, 2006 7:36 am
				by diana
				What about centrigation? If it is not possible to precipitate, I centrifuge my samples (5min at 4000 upm should be enough) and afterwards I use a 0.45 µ filter.
			 
			
					
				
				Posted: Mon Nov 27, 2006 12:42 pm
				by Peter Apps
				Hi Mattias
This is going to be tricky - colloidal silicas have particles going down to molecular dimensions, and depending on the surface treatment they are macroscopically sticky.
The only thing that I can think of is to spin them through a very high MW cut off ultrafilter - the kind that is used to take proteins and other large molecules out of biological samples.
Good luck    
Peter
			 
			
					
				
				Posted: Mon Nov 27, 2006 6:57 pm
				by Mark Tracy
				You can also try Anotop+ 0.02µm filters from Whatman; that's right 20nm. They also have a glass-fiber prefilter built in.
			 
			
					
				
				Posted: Tue Nov 28, 2006 8:29 am
				by HW Mueller
				I have a secondary interest in this, sort of. We have had horrendous trouble with  transition metal oxides (for instance Y oxides stick to everything).  From synthetic organic chem.  there are still these ugly memories of Al oxides. All can and could be handled after acidification, so I wonder if Uwe´s suggestion is not the way to go?
			 
			
					
				
				Posted: Tue Nov 28, 2006 11:52 am
				by Mattias
				I added some HCl and it worked!
The silica precipitated and floated on the surface. It was possible to filter the solution with the syringe filter (0.22 µm). Now I have to see what happens with my analyte...
			 
			
					
				
				Posted: Tue Nov 28, 2006 3:18 pm
				by HW Mueller
				Good to know that Si oxides are similar to the others.
			 
			
					
				
				Posted: Wed Nov 29, 2006 6:39 am
				by Mattias
				Of course my analyte was hydrolysed as well....
Removal of Aerosil from samples must be a very common problem, since it is used in many solid pharmaceutical formulations.
Doesn't winemakers use something to remove particles from wine?
			 
			
					
				
				Posted: Wed Nov 29, 2006 7:37 am
				by HW Mueller
				Maybe a tight pH, time, and temp. control will do the trick? Or ionic strength changes might modify the gel so that it can be filtered?
			 
			
					
				
				Posted: Thu Nov 30, 2006 6:49 am
				by Mattias
				The situation has been solved! When I increased the volume of solvent, the solution could be filtrated much more easily.
I then injected 1000 µl (instead of 100 µl) on a small precolumn. By using a column switch I then eluted my peaks to the analytical column. I probably still inject particles, but it will only kill my precolumn. I see no problem the first 10 injections at least.