Page 1 of 1

GCMS Saturn 2000

Posted: Tue Oct 03, 2017 7:55 am
by Jihad
Hi all,
One of my GC ms sensitivity is very low. 10 ppm organophosforus std its giving 900khz only for 1ul injection.Splitless injection .
In diagnostics RF is failing saying like RF DAC cannot convert value above 56 .software suggesting RF generator board faulty .replaced RF generator board but its giving same message in diagnostics. I also changed main board
but there was no progress. Air and water is normal . calibration is passing . trap calibration is failing..please advise .

Re: GCMS Saturn 2000

Posted: Tue Oct 03, 2017 9:55 pm
by James_Ball