Method re-validation criteria
Posted: Mon Oct 02, 2017 9:05 pm
by umoshplc
Hello All,
I got a strange situation with my method, which was originally validated and submitted to regulatory. Now with the same method samples are assaying about 95% and I need to increase 5% of the organic content of the sample diluent to get 100% assay.
Do I need to do full validation for this change? where can I find the guidance related to this? Please advise if you know. Thanks.
Re: Method re-validation criteria
Posted: Tue Oct 03, 2017 4:59 pm
by tom jupille
The general rule is that modifications to a method require full revalidation. Certain de facto adjustments are permitted by ORA Lab 5.4.5, but changes to sample prep are not included. Bottom line: yes, you have to validate the modified method (sorry!).
Posted: Tue Oct 03, 2017 7:52 pm
by umoshplc
Thanks Tom,
this change what I suggested does not fall under the scope of ORA 5.4.5, do you think I still need to do full validation.
Posted: Wed Oct 04, 2017 2:40 am
by tom jupille
Yes, that's the whole point. Changes within the scope of ORA Lab 5.4.5 are "adjustments" and do not require revalidation so long as you meet system suitability. Anything else is a "modification" and *does* require revalidation.