Negative y intercept LCMS

Posted: Sat Sep 30, 2017 7:39 am
by SonicW
Hi all,
What are some possible causes of a negative y intercept (i.e. a positive x intercept) in a calibration curve (ESI LCMS method)?
The standards seem to have been prepared correctly (they're MeOH based standards spiked into a blank urine matrix).

Thanks for your help :)

Re: Negative y intercept LCMS

Posted: Sat Sep 30, 2017 10:40 am
by Peter Apps
Welcome to the forum.

This is a hardy perennial. Possible causes include non-linearity, adsorbtive (or other) losses of analyte, a signal threshold set too high, not taking the purity of the standard into account, poor repeatability, .....

What your particular problem is will depend on the details of your analysis - tell us those and we will be able to help.

A search of the archives will throw up plenty to keep you busy also.

Peter

Re: Negative y intercept LCMS

Posted: Sat Sep 30, 2017 8:54 pm
by SonicW
Hi Peter, thanks for your reply.
So the issue is with Pholcodine. I use a mass-labelled codeine as internal standard.
I'll double check the purity when I'm in the lab next week.
For adsorbtive and other losses- do you think degradation of analyte in the standard master is likely? While codeine is reasonably stable in MeOH, I'm not sure about Pholcodine.

Thanks for your help.