I´m injecting 10-20 µL. I suspect column is not overloaded. I´m using DAD.
As far as overload is concerned, what matters is not the injection volume so much as the mass injected.
Saying you used a DAD doesn't tell much. What wavelength. If you look at the structure of clindamycin, there is no much in the way of a UV chromophore, which means that you will probably be limited to the "end absorbance" of carbonyls down around 210 nm -- and probably won't have a very high molar absorbance even there.
All of which makes me suspect that if you are injecting enough material to see a peak, you may well be overloading your column.
If this were my problem, the first thing I would do is to obtain a UV spectrum of clindamycin in something approximating your mobile phase. That will let you establish the absorbance maximum and at least estimate the molar absorbance. Then go and do a loading study -- inject your standard, then dilute by a factor of 5 or so and reinject. If the peak shape improves, then you were overloading. If it doesn't improve, that will confirm that overload is not an issue.