Help! Using Nitrogen carrier gas in GC-MS Application!

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
We are dealing with the nationwide issue of low helium supplies. I switched over to UHP nitrogen in my Varian 3800 GC - MS (2000) yesterday and pumped down overnight. In the column parameters, I made sure to switch the inlet I'm using from Helium to Nitrogen. Is there anything else I need to do? The inlet is maintaining pressure well but I certainly can't pass an air/water check, and I don't think the the problem is a leak. When I attempt an autotune, it can barely detect 69 m/z and all the calibration masses are way off calibration (i.e. it's detecting 131 at 135, etc). Any suggestions for using nitrogen as a GC-MS carrier gas?

-Aaron

Why not try H2?

N2 can be ionized at 15.5 eV, which is a lot lower than that of He(24.8 eV) and H2(18eV). That can cause a lot of problems when using EI. This topic had been talked about in the forum.
2 posts Page 1 of 1

Who is online

In total there are 13 users online :: 0 registered, 0 hidden and 13 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 13 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry