comparison of impurities during site method transfer

Posted: Sun Nov 19, 2006 10:16 pm
by srdales
What criteria are individuals using for a site transfer on an impurity method?

Differences of 30%, or 50% if you are comparing impurities less the 0.1%?

Posted: Mon Nov 20, 2006 5:18 am
by JM
We use RSD of 15 % for impurities above 0.1 % and above LOQ. We do not consider impurities below 0.1 % during method transfer.

JM