%RSD issue using an Internal standard
Posted: Fri Sep 22, 2017 6:09 pm
Hello all, I have a question regarding the %RSD of GC standards when using the IS. We are performing the USP method for chlorobutanol assay, the internal standard is 15 mg/mL 2,2,2-trichloroethanol. Chlorobutanol std conc. is 10 mg/mL. USP has a specification for %RSD of chlorobutanol/IS ratio of NMT 0.3%. We are having a hard time achieving the spec, the best time we got 0.36%. The majority of time we got around 1%.
Because it is USP method, there is not much room we can change. We contacted FDA, and was told they can easily achieve the required %RSD (0.04%-0.13%). Could you please provide some suggestion of how to improve the reproducibility? Any suggestion is welcome.
Here is the condition:
Injection port: 260°
Detector: 280°
Column: 135°
Carrier gas: Helium
Flow rate: 1.0 mL/min
Injection volume: 1 µL (use 10 uL syringe)
Injection type: Split injection, split ratio 10:1
Run time: 12 min
column is 0.32-mm × 30-m fused silica, 0.25-µm DB-wax (brand new column)
liner: Restek split liner with wool
Septum purge flow: 3 mL/min
injection speed: 500 uL/sec
Thank you!
Because it is USP method, there is not much room we can change. We contacted FDA, and was told they can easily achieve the required %RSD (0.04%-0.13%). Could you please provide some suggestion of how to improve the reproducibility? Any suggestion is welcome.
Here is the condition:
Injection port: 260°
Detector: 280°
Column: 135°
Carrier gas: Helium
Flow rate: 1.0 mL/min
Injection volume: 1 µL (use 10 uL syringe)
Injection type: Split injection, split ratio 10:1
Run time: 12 min
column is 0.32-mm × 30-m fused silica, 0.25-µm DB-wax (brand new column)
liner: Restek split liner with wool
Septum purge flow: 3 mL/min
injection speed: 500 uL/sec
Thank you!