flow H2 and air goes down en up

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Hi,
I am using a 6890 Agilent GC. Currently the flow of H2 is 40ml/min and air flow 450ml/min.
The instrument is ready if this is OK. Normaly it is ready and after 2min is again not ready. The flow goes down en up after 2min (ready and not ready).What could be happening. My guess was JET replace. I made it and after this it is again the same.

Please help me. Thanks

Check that the pressure of air and hydrogen in the supply to the GC is not fluctuating.

Is it only the hydrogen that fluctuates, or the air and the hydrogen ? If both I suspect an electronics or hardware problem.

Check on the software, presumably Chemstation that you are not accidentally programming the flows, or turning the detecotr on and off.

Peter
Peter Apps

Could it be that your flame is going out? When this happens, the software will decrease the gas flows and then increase them as it tries to re-light.

Hi,
thenk jou everybody!!! I found the problem it was a lit offset. I put 1 and everything is ok now
4 posts Page 1 of 1

Who is online

In total there are 17 users online :: 2 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Bing [Bot], Google [Bot] and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry