Buffer gradients/ IP gradients

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
My question is in regards to buffers and running gradients. For example, if I will prepare a 10mM phosphate buffer for reservoir A and want to run a gradient (5%-95%B), would I need to add phosphate to reservoir B, the organic (e.g. ACN/MeOH), and thereby make it 10mM as well? I assume not, but don't know.

If I don't add phosphate buffer to the organic reservoir (B), then what would one consider the optimal buffer concentration if peaks eluted ideally, for example, at 50% A(10mM) and 50% B? In other words, would an isocratic run to follow use a 5mM phosphate concentration for mobile phase (50:50)? Or would I keep the phosphate concentration at 10mM? This probably sounds elementary, but hope to get some feedback! Thanks alot!

Lastly, I heard one should not run gradients with IP reagents such as alkylsulfonates. Any thoughts?

Thanks,
Phil
Thank you for your response. It is appreciated!

I think it is a good question, which I heard the first time.
Here is my thinking:
1) usually we want enough buffer capacity instead of accurate. for example in your gradient
For example, if I will prepare a 10mM phosphate buffer for reservoir A and want to run a gradient (5%-95%B),
at 95% B MP your buffer concentration is 0.5mM. If 0.5mM is enough, the whole run has good buffer.

2) if you want to keep the buffer concentration constant through the gredient you can (if you are able to since you may not be able to dissolve the same amount of the salt into the organic). this may help a bit for the base line if you do though may not much.

the major reason preventing IP gradient run probably is that it takes a long time for the column to stablize in the IP MP. an unstabilized run gives rough base line.
Excel

Thank you for the feedback ym3142. I greatly appreciate the insight. Certainly, the question wasn't worded the best, but it sounds like you caught on to what I was attempting to communicate.

Basically, it would seem then that the optimal isocratic follow up would then be (potentially) the decreased mM concentration relative to the % organic ramped in... True, the salt solubility would not be the same as that of the organic, but sounds like it helped to get the string of ideas across.


Thanks alot for the feedback,

Phil
Thank you for your response. It is appreciated!

When I am doing development work, I often use ternary systems with organic in A, water in B, and buffer/IP/etc in C. I can always turn it into a binary system later.
Mark Tracy
Senior Chemist
Dionex Corp.
4 posts Page 1 of 1

Who is online

In total there are 20 users online :: 3 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot], Google [Bot] and 17 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry