dichloroethylene detection with GC/FID
Posted: Mon Nov 06, 2006 11:05 pm
Hi
I am trying to develop a calibration curve for cis-dichloroethylene using GC/FID. But I haven't been able to see any peak so far.
I have a stock solution of cis-DCE (0.85mg/ml) in methanol and inject 0.005 of the stock solution in a 10ml vial containing 5ml of water. This gives me a concentration of 0.85 mg/L in water and approximately 0.1 mg/L in gas. After one hour I take a 25microliter headspace sample and inject it in the GC.
I am running the samples in a GC with a Rt-QPLOT column (from restek). The oven temperature that I am using is 60 degree C hold 2 min, 10 degreeC/min to 150 C hold 5 min. The detector temperature is at 250 C. The injector at 200 C.
My idea is that I am injecting a small volume to the GC. The lab manager recommended me to inject at most 25uL. Do you have an idea on what could be the reason for such a small volume of injection? Do you have any idea of what is wrong with my method? Do you have some suggestions to improve the method?
THANKS
ITZA
I am trying to develop a calibration curve for cis-dichloroethylene using GC/FID. But I haven't been able to see any peak so far.
I have a stock solution of cis-DCE (0.85mg/ml) in methanol and inject 0.005 of the stock solution in a 10ml vial containing 5ml of water. This gives me a concentration of 0.85 mg/L in water and approximately 0.1 mg/L in gas. After one hour I take a 25microliter headspace sample and inject it in the GC.
I am running the samples in a GC with a Rt-QPLOT column (from restek). The oven temperature that I am using is 60 degree C hold 2 min, 10 degreeC/min to 150 C hold 5 min. The detector temperature is at 250 C. The injector at 200 C.
My idea is that I am injecting a small volume to the GC. The lab manager recommended me to inject at most 25uL. Do you have an idea on what could be the reason for such a small volume of injection? Do you have any idea of what is wrong with my method? Do you have some suggestions to improve the method?
THANKS