Hplc troubleshooting
Posted: Wed Sep 13, 2017 1:49 pm
by shiva baskar
Hi professionals.
My question is very general. Do pump check valves need to be sonicated or just flushed by using syringe.
Few service engineers suggest me to sonicate few suggest to flush with syringe.
Please help me in this regard.
Re: Hplc troubleshooting
Posted: Wed Sep 13, 2017 2:09 pm
by HPLC chemist
Both, as long as it works. I generally flush 1st (Water, Methanol, Acetonitrile, IPA, Hexane..) with the column offline. If that doesn't work I extract and sonicate (after a visual examination).
Re: Hplc troubleshooting
Posted: Thu Sep 14, 2017 4:29 pm
by DR
If you are using ruby/sapphire check valves and pumping straight ACN, you can avoid the need for flushing/sonicating by switching to ceramic.
Re: Hplc troubleshooting
Posted: Thu Sep 14, 2017 4:48 pm
by shiva baskar
Thank you sir. But I am using ruby check valves how can I identify when the life time of it is completed.
Re: Hplc troubleshooting
Posted: Thu Sep 14, 2017 7:46 pm
by HPLC chemist
When they are scratched, warped, full of gunk.
Re: Hplc troubleshooting
Posted: Fri Sep 15, 2017 4:43 pm
by DR
...or annually, as part of a routine PM service (more frequently if the system is used lots or problems come up regularly between PMs).
Re: Hplc troubleshooting
Posted: Sun Sep 17, 2017 11:15 am
by BostonFSE
you can remove the check valve and hold them up to your ear while shaking it.
If its good, you will hear the tiny ruby ball make a rattling sound ( the ball going up and down)
If you do not hear this sound, it is stuck and should be sonicated in IPA or meoh.
Of course, when you have consistent pressure fluctuations from your trace- the main culprit is check valves, then pump seals.
Re: Hplc troubleshooting
Posted: Sun Sep 17, 2017 11:28 am
by shiva baskar
Thank you sir. This will help me alot in troubleshooting