My problem is that they ship the column in Hexane + 3% Acetonitrile. Before I started my analysis I flushed as follows:
Hexane/THF
THF
THF/Acetonitrile
Acetonitrile
Mobile phase listed above.
I know the performance of these silica amino columns deteriorates because of the pH problem, but I have used this column once before a couple of years ago and got good separation between rhanmose, fucose, xylose, arabinose, mannose, glucose, and galactose. However, the last two times I have purchased this column, the retention factor and resolution on mannose, glucose, and galactose were much worse on my initial injections.
I also have the Shodex Asahipak, and Phenomenex Luna amino which are pH stable, but neither of those separates all of the sugars that have the same MW under MS friendly conditions.
So, my question is, how would the flush procedure damage the column? I thought water would be the only solvent to damage it.