problem with the sample peaks

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

4 posts Page 1 of 1
Although I get the proper peaks for standards I couldnt get the suitable peak with the samples. I am trying the determine the resveratrol contents of the red-wines. I used SPE (c18 cartridges) technique in the sample preparation step.
Chromotographic conditions:
Column: ODS c18 (250x4.6mm id, 5 um)
Detector: UV (306 nm)
Loop vol: 20 mL
Temp: 25 degree centrigrad
Mobile phase: ACN/Water/Acetic Acid (24.9/75/0.1)
Flow rate: 1 mL/min.
I also send the photos of standard and wine-sample peaks. If you have suggestions I will be glad.

Image

Image

how did you prepare the standard and the sample? were their composites very close to the mobile phase? if not, try so.
Excel

I would attribute such things that you see to a very dirty mobile phase. Are you recycling the mobile phase?

It is looking that there is a problem with the mobile phase. have you injected the mobile phase or not and have you checked the individual components for inerfering. if so try with different diluents so that you will get good shape.

g.r.reddy.
4 posts Page 1 of 1

Who is online

In total there are 29 users online :: 2 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry