if the chemicals built up on the column the retention time is supposed to be longer.
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Actually, retention usually decreases as a result of chemical contamination. The junk ties up part of the stationary phase surface and therefore changes the phase ratio. Only if your analytes interact more strongly with the junk than they do with the original stationary phase will retention increase.f the chemicals built up on the column the retention time is supposed to be longer.
Are you running a gradient ?
Often with a gradient method the column never reaches perfect equilibrium with the starting conditions (there was a paper on this in LCGC a while ago, but dang if I can find it...), but does reach a semi equilibraitation that's the same every time, so the first few runs the equilibriation is changing from your pure starting conditions to the gradient.
Try running a few blanks before your run and see if that helps.
Paul.
I put some TTBB into the solution. Since the retention time for TTBB in every run is the same, so can say it is not the problem of the pumpWhy rule out the pump?
if that is the case what solvent can I use to flush the column?Actually, retention usually decreases as a result of chemical contamination. The junk ties up part of the stationary phase surface and therefore changes the phase ratio. Only if your analytes interact more strongly with the junk than they do with the original stationary phase will retention increase.f the chemicals built up on the column the retention time is supposed to be longer.
We worry that people forget to tell us details, and you thought that your system is using CO2 SFC with a modifier isn't important?. I assume the method used to work OK, but if it didn't - then then sample introduction method should be reviewed to ensure consistent dissolution.actually i am doing supercritical fluid chromatography. Thus the mobile phase is supercritical CO2 with modifier( Methanol V:V=0.18:1). My sample is Flurbiprofen (dissolved in MeOH) The peak shifting I mentioned appeared during a series of injections.
How come my peaks oscilate ? First move forward and then backward. They are all continous injections under same method.