Rolando » Thu Sep 14, 2017 2:34 pm
What is your:
- column head pressure
- transfer line pressure
- vial pressure
- split ratio
- oven temp program
Also,
- What are your analytes and matrix?
Also - check the top of the transfer line needle after pushing it through the septum to attach to the inlet. Sometimes a bit of septum gets stuck in the very top of the needle.
I have tried different pressure settings, what would be the correct pressures and split ratio.
The analytes are: vinyl chloride, 1,1-diclhoroethylene, chloroform, carbon tetrachloride, 1,2-dichloroethane, trichloroethylene and tetrachloroethylene.
The oven ramp is init.30°C 4.5min, 30°C/min to 135°C 4min and 10°C/min to 220°C.
Inlet temp is 175°C
Column head pressure 20