Hello all,

I'm trying to use USP method for API verification. It's a gradient one and takes 72.1 minutes. Analytical column is 4.6-mm 25-cm; 3-mm packing L1, MP-A is 10 mM ammonium acetate and 5 mM sodium 1-heptane sulfonate, and MP-B is Acetonitrile and methanol mixture. The column and the system were equilibrated for 3 hours. At the beginning, we've got perfect peaks -Tailing factor: NMT 1.5 for both peaks-. However, after 10-15 injections, we lost system suitability; unacceptable peak shape. The column is new and used only for this analyze. Since it's USP method, we have to use it. Is there any suggestion? Regards.

Re: Ion-pair and peak shape...

Posted: Tue Sep 05, 2017 4:34 pm
by tom jupille
More information would help:
- did the peak shape deteriorate gradually or abruptly?
- "peak shape" is a bit vague; is it simply tailing or something else (e.g., a shoulder0
- are all the peaks in the separation showing the same problem (or are some peaks still well-formed)?
- was the problem accompanied by changes in retention time?

The reason for asking all of this: if it's a "physical" problem (column headspace or partially-blocked frit) is should affect all the peaks in the separation in a similar manner, and the retention times should be unchanged. If it's a headspace, it will usually come on gradually; a partially-plugged frit will usually come on abruptly.

If it's a "chemical" problem (e.g., deterioration of the bonded phase), it is likely to affect some peaks more than others, and the deterioration is usually gradual.

Re: Ion-pair and peak shape...

Posted: Tue Sep 05, 2017 5:57 pm
by bircanyaman
Dear Tom,

Thank you for reply.

- The peak shape deteriorated gradually
- mostly shoulder
- There're five peaks in the chromatogram ( main peak -R+S-, İmp A,B,C,D and E. We observed all the same problem for all peaks
- There were no change in retention times

As you explained bellow, it should be physical problem. Using pre-column filter could resolve problem. We'll try and let you know result. Thanks :)

Also, we observed that there was precipitate in MP-A (ammonium acetate probably). Is there any tips to prepare/store MP contain ammonium acetate? We prepare MP daily and filter from 0.45 micron filter.
More information would help:
- did the peak shape deteriorate gradually or abruptly?
- "peak shape" is a bit vague; is it simply tailing or something else (e.g., a shoulder0
- are all the peaks in the separation showing the same problem (or are some peaks still well-formed)?
- was the problem accompanied by changes in retention time?

The reason for asking all of this: if it's a "physical" problem (column headspace or partially-blocked frit) is should affect all the peaks in the separation in a similar manner, and the retention times should be unchanged. If it's a headspace, it will usually come on gradually; a partially-plugged frit will usually come on abruptly.

If it's a "chemical" problem (e.g., deterioration of the bonded phase), it is likely to affect some peaks more than others, and the deterioration is usually gradual.