Alkylated PAH analysis
Posted: Fri Oct 27, 2006 5:02 pm
by Sean
I have a mixture of PAH in an alkane solution that I am having trouble separating with ordinary methods. Most of the PAH in this solution have a high degree of alkylation making them extremely difficult to resolve (there are a very large number of PAH, and they tend to coelute).
My understanding is that a normal phase column will separate PAH by ring number with little sensitivity towards the alkyl groups. So my plan is to use a normal phase column to fractionate the sample by ring number, then use a reverse phase column to analyze each fraction.
I plan to buy a cyanopropyl column for this purpose, but I am wondering if there is a stationary phase better suited for this purpose. Is there a column that is more sensitive to ring number and less to alkyl groups?
Thanks in advance.
Posted: Fri Oct 27, 2006 5:22 pm
by Uwe Neue
It is not normal phase per se that tends to separate the PAH according to ring number. You need an amino column to do that.
Posted: Fri Oct 27, 2006 6:11 pm
by Sean
Thanks for the clarification. I was aware that I needed a specific stationary phase, but I didn't make it clear in the post. I believe my choices are either a cyano or amino column.
I had heard that an amino column would be ideal, but I am also told that compounds with carbonyl groups (which are present in my sample) will bind permanantly to the stationary phase. Is this true?
Posted: Fri Oct 27, 2006 9:32 pm
by Uwe Neue
I am not aware that cyano columns do this group separation as well. I have some doubts though. Nitro-phases do this though via clear pi-pi interactions.
You are right that amino phases are prone to react with aldehydes and ketones. Since you have such things in your sample, the amino column is not a good idea.
I would go for the nitro-phase, but you may have to dig until you find a supplier. Check with Macherey-Nagel in Germany!
Posted: Mon Oct 30, 2006 10:54 am
by HW Mueller
How do phenyl phases or Hypercarb behave with PAH?
Posted: Mon Oct 30, 2006 9:37 pm
by Uwe Neue
Phenyl columns give more retention for pi acceptors, but I don't think that this will help here.