CPR for Waters Xterra C8 column?

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
Hi Members,
What would you do if impurity peaks starting to merge with the main peak? (the initial resolution between impurity and main peak is about 1.5) column: Waters xterrra C8, mobile phase H2O and ACN, gradient 90% water to 10% water after 40 minutes. The main compound is testosterone palmitate and the impurity is a little tiny unknown peak. I noticed the same behavior in two "new" Waters C8 columns (about a month old). I had no choice but to buy another two columns to complete the validation work.
Thanks in advance,
ntruong
It appears that something in your sample matrix is tying up the active sites in your column. Without more information, it is difficult to say exactly what is happening. You can try to clean the columns up by following the manufactirers cleaning procedure that came with the column.
George Balock

Tough to tell. The sample matrix is a possibility. Also for such small changes the column temperature can be relevant. Also, consider going back to store the column in the same solvent in which it came (acetonitrile).
3 posts Page 1 of 1

Who is online

In total there are 29 users online :: 2 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry