The intention is perform a total validation from a modified existent method (USP), sample is neat API.
New method would change mobile-phase and column type (ID, length and particle size)
1-Calibration curve 5 points-duplicate each---------10
2-Limit of detection and quantification dilutions---------5
3-Upper limit of detection (also for Range)125,150%- duplicate---------4
4-Recovery (accuracy) in 2 concentrations 50, 100% of theory---------5
5-Precision for sample %RSD 3 conc/3 replicates each---------9
6-Ruggedness- other analyst performing calibration
(5 points, recovery 2 spikes, reproducibility 6 replicates)---------14
7-Robustness- %B variation of + and - 5% , flow rate
variation + and – 20%---------4
Number of runs---------51
Of course, some runs are made to extrapolate a possible detection limit before do the dilutions.
I really don’t think that step 6 is necessary, since we have results of standard method that we can do a T test for assure equivalence but people insisted.
Would you add/remove something?
Fabiano