I have two 5890s on the same breaker, one has the baseline interference, the other does not. The first thing I did was run a dedicated line to the gc that had the interference. No change in the interference.
I recently installed a zero air generator, I bypassed it using compressed air, my baseline rose and the interference stayed the same.
I leak checked the system, and checked flow rates, and everything looked ok.
Finally I turned on the other 5890 lit the fid, and noticed the interference on the other gc went away. pic below

I checked the outlets and all are grounded correctly. The two gcs are on separate breakers now, but the computers for both gcs are still on the same breaker as the gc that didn't have any interference problems to begin with. I'm going make sure that the computers are both on the same breakers as the gcs they are connected to, if that doesnt work im at a bit of a loss.
Any ideas? Having both gc's on and lit fixes the problem, but it'll drive me nuts not knowing why that is.