High temperature septum recommendation needed for GCMS
Posted: Mon Oct 23, 2006 10:43 am
We have a Thermo Trace GC 2000 fited with a SSL injector that we are running at 350°C. This high temperature is necessary to measure some high-boilers in our samples by GCMS. At the moment we are using Thermogreen LB-2 septa, which are rated to 350°C. However, we find that at any SSL temperature above 310°C, the septa start to melt and crumble at the injection point, resulting in a massive increase in bleed/background.
The disparity between the septum rating and what actually happens at the SSL set temperature may be because the Thermo SSLs are known to heat the liners more "evenly" than other brands. This means the septa are heated more. The septum rating then depends on what SSL it was tested on.
Regardless of the niceties of how the LB-2 was rated, we still need a better performing septum for our method. Can anyone recommend another brand of septum that will perform better than the LB2 at 350°? This has to be compatible with the Thermo SSL - i.e. 17mm diameter.
thanks
Tony
thanks
Tony