I just bought a new Chromolith RP-18 endcapped LC column and used it for separation of steroids and steroid conjugates. The column is fairly new (3 month) but over the last month peaks were getting more tailed and now I have shoulders on them. I regualarly washed the buffer away with water and stored it in 60/40 ACN/water. However, it seems that even though I'm using inline filter and a guard column the column head is somwhow blocked as I have an unusual backpressure of about 410 psi (with 100 % ACN at flow rate 2.0 ml/min). It should be about 190. I already run the complete washing procedure with water-ACN-n-propanol-ACN-water.
The shouldering is getting better after having about 1 L of water passed over the column but has not yet dissappeared. I just wonder were the contaminats came from since I use all prodection and washing. Anyone with a bright idea pls answer!!
Kind regards
Frank