Hi!I'm new to operate GC-MS, I'm in trouble to situation like photos in the link

https://drive.google.com/drive/folders/ ... sp=sharing

I'm trying to measure several testing blank , and as you can see, the result of area of compound PAD is really low, all of them are under 100(large data in the middle are standard solution, 4 ppm and 12 ppm).
I've got two comment to this problem:one said that I didn't setup right so machine cannot catch the right m/z ion, and the other said I should raise sensitivity.
To the first comment, I use SIM to setup when to start to detect the ion, and setup its target ion, I don't understand what more I can do?And to the second comment, I'm sure all consumable are new and clean, what else I can do to improve sensitivity?

The model of GC/MS:Agilent 7890B/5977B
Columns:DB-5ms Ultra Inert 30 m X 250 um X0.25 um
Welcome to the forum.

The peaks in a blank should be very small. You only have a problem if the peaks from standards are very small. Your peaks from standards are not very small, and so you do not have a problem.

Peter
Peter Apps
If it truly is a blank sample then Peter is correct, you should have little or no peaks in a blank, unless it is contaminated. Unless it is an internal standard you are looking for, then the peaks should be the same size in all analysis.
The past is there to guide us into the future, not to dwell in.
3 posts Page 1 of 1

Who is online

In total there are 10 users online :: 0 registered, 0 hidden and 10 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 10 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry