Quadrupole MS cleaning

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

3 posts Page 1 of 1
I would appreciate if anyone can suggest me how to clean Quadrupole metal rods and how to take them apart. I have Unicam's Nermag Automass (610/615) instrument and I don't have its manual. If anyone has manual for this/similar model that will be very helpful.

Thanks.

With the Agilent system if you take apart the quad rods I believe you then have scrap!
The alignment of the rods is crytical to the operation of instrument so I would be tempted to get in touch with the manufacturer to gain some guidance.
The gcms systems I use have never had the rods cleaned and they work ok, the source however does require cleaning and this is a relatively simple task.

GCguy
GCguy

Some of our instruments never need the quadrupoles cleaned under normal circumstances.

Some have prequads that are cleaned.

However, the Finnigan TSQ70/700 which we purchased in 1988 needs the quads cleaned every one to two years. We take the quads apart and clean lightly with an abrasive such as 3m paper or Ajax with a Qtip. Then wash them very well with water, then methanol or acetone. Then blow off with nitrogen or some other clean/nontoxic gas.

Then usually hold up to light at different angles to make sure they are clean of all residue. Finnigan supplies a jig to allign the quadrupole. However, you need to be very careful in how much you tighten the screws in the insulating collars on the quadrupoles. Think you tighten all the way then loosen a 1/2 turn? or something like that.

For the very first years on the Finnigan, we just cleaned the ends of the quads closer to the source with an abrasive with either the Q-tip cotton tip or by breaking the Q-tip stick and using abrasive on the end of the stick. Most of the burn noted on end of quad since no prequad with this system. We had very long Q-tips which we used I think with methanol to wash the rest of the Quad. Then backflush the quadrupole assembly thoroughly with water from the faucet. Hold up in light and make sure no abrasive film noted, then wash with methanol or acetone to dry the rod. Think we put in the GC oven 1t 50 C lighly wrapped in tinfoil for 5-10 minutes to dry, then blow off with gas.

Either method seemed to work fine, the latter probably involves much less risk. Have probably done the former procedure about 10 times and the latter one about 4 times.

If you take the QUads apart, be sure to note which rod was in which holder, and place back exactly the same. I think our quad rods had serial numbers on the end of the rods.

Instrument still running routinely today, replacing soon with another instrument, but work reasonably well doing routine GCMS EI and CI.

Dan Gartiez use to have some Nermag instrumentation, maybe you could contact him for some tips..

http://www.ghg.net/texms/
Sailor
3 posts Page 1 of 1

Who is online

In total there are 17 users online :: 1 registered, 0 hidden and 16 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: Semrush [Bot] and 16 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry