HP 5890 II Noise
Posted: Thu Aug 10, 2017 4:58 pm
I am running a modified version of ASTM D6729 on a refurbished HP 5890 II Plus GC/FID with a HP 7673 autosampler using the following conditions:
Injection: Split, 150:1
Volume: 0.5 uL
Oven ramp:
Initial: 30C
Ramp 1: 1.00 C/min to 50C
Ramp 2: 0.5 C/min to 75C
Ramp 3: 30C/min to 270C
Column: Restek Rtx-DHA fused silica column - crossbond 100% dimethyl polysiloxane. L: 100m (has been cut to ~90m), ID: 0.25mm, df: 0.5um
Head pressure: 37.8psi, constant flow mode
Carrier: He
I am analyzing samples from a liquid hydrogenation reactor loaded with Pd catalyst running a synthetic feed of 5% 1-octene and 1% styrene in hexane solvent. The hydrogenation products are n-octane, ethylbenzene, ethylcyclohexane, and 5-6 octene isomers.
I have two issues (which may or may not be related) with this instrument:
1) the % RSD will be 4-6% for the analytes of interest. This is from repeated injections from separate vials of the same standard. I have seen better (1% or less %RSD) in the past so I know the instrument and the method are capable of better.
2) Excessive noise/spikes on the chromatogram as seen on the picture below. These two samples were injected within 25 hours a part as part of a continuous run of 30+ samples. The noise will come on and dissipate at random times. I have also observed the noise while opening the oven door while the FID was OFF. I have also observed it right after an injection while the autosampler turret was moving, however, this doesn't happen on every injection. I am almost certain the problem is electronic in nature. I do not know how to track down the source.
I have changed the EPC, selenoid, the entire inlet assembly (including split vent line), main board, the entire FID assembly including the heater, PRT, and FID board. I am running out of things to swap out! I do not have a spare EPC board or cable and am finding it difficult to convince myself that would make a difference.
Any wisdom in this area would be appreciated. I feel like i could track down the source if I was better with a multimeter ?
imgur link in case the above image does not display: http://imgur.com/MUAjAw9
Injection: Split, 150:1
Volume: 0.5 uL
Oven ramp:
Initial: 30C
Ramp 1: 1.00 C/min to 50C
Ramp 2: 0.5 C/min to 75C
Ramp 3: 30C/min to 270C
Column: Restek Rtx-DHA fused silica column - crossbond 100% dimethyl polysiloxane. L: 100m (has been cut to ~90m), ID: 0.25mm, df: 0.5um
Head pressure: 37.8psi, constant flow mode
Carrier: He
I am analyzing samples from a liquid hydrogenation reactor loaded with Pd catalyst running a synthetic feed of 5% 1-octene and 1% styrene in hexane solvent. The hydrogenation products are n-octane, ethylbenzene, ethylcyclohexane, and 5-6 octene isomers.
I have two issues (which may or may not be related) with this instrument:
1) the % RSD will be 4-6% for the analytes of interest. This is from repeated injections from separate vials of the same standard. I have seen better (1% or less %RSD) in the past so I know the instrument and the method are capable of better.
2) Excessive noise/spikes on the chromatogram as seen on the picture below. These two samples were injected within 25 hours a part as part of a continuous run of 30+ samples. The noise will come on and dissipate at random times. I have also observed the noise while opening the oven door while the FID was OFF. I have also observed it right after an injection while the autosampler turret was moving, however, this doesn't happen on every injection. I am almost certain the problem is electronic in nature. I do not know how to track down the source.
I have changed the EPC, selenoid, the entire inlet assembly (including split vent line), main board, the entire FID assembly including the heater, PRT, and FID board. I am running out of things to swap out! I do not have a spare EPC board or cable and am finding it difficult to convince myself that would make a difference.
Any wisdom in this area would be appreciated. I feel like i could track down the source if I was better with a multimeter ?
imgur link in case the above image does not display: http://imgur.com/MUAjAw9