IC columns and mobile phase compatibility

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
I´m looking for some information about mobile phase compatibility for IC anion columns, I have found that some silica based IC columns are not compatibles with Methanol or Acetonitrile (Vydac or Dionex for example) while others (alltech) can run up to 100% organic.
Is this due to silica modifications?

regards

K4

I doubt that the Dionex IC columns are silica based. Solvent compatibility depends on how the ion-exchanger groups are attached to the core particle. These groups can be attached covalently (e.g.Hamilton PRP-X100) or a ion-exchange latex is attached to a sulfonated core particle through electrostatic interaction (most Dionex columns).
It is possible to strip the latex from the substrate. This is not possible with covalently attached ion-exchange groups.

Always refer to the column manual for solvent compatibilty information.
--
Robert Haefele

thanks Robert,
You are rigth Dionex is not based on silica,
thanks for the information on Hamilton PRP

K4
3 posts Page 1 of 1

Who is online

In total there are 21 users online :: 0 registered, 0 hidden and 21 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 21 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry