Anybody who are using gas chromatography 7890A help me. I am using GC7890A with cp-wax 58 ffap 25m length 0.53mm diameter column. But it is not working tried so many times stilll flame out. Once light out and inject one or two sample then flame out. What i do help mee plzzz
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
Don't double post !
-
- Posts: 61
- Joined: Thu Jun 14, 2012 6:33 pm
Are you injecting water as your solvent? If so, try reducing the injection volume as a first port of call.
-
- Posts: 3477
- Joined: Mon Jan 07, 2013 8:54 pm
Need to know injection solvent, column flow rates, gas flow rates in FID and any other information we can get to help.
The past is there to guide us into the future, not to dwell in.
-
- Posts: 3
- Joined: Fri Aug 11, 2017 2:22 pm
Sorry for double posting. I really need help
Injection volume 0.1ul, pressure 0.8 psi, total flow-183 Ml/min, septum purge flow-1ml/min, split ratio 150:1, air-200 ml/min, h2-20ml/min, he- 15ml/min.
Washing solvent-ethanol .
Injection solvent- benzene and isooctane standard
Injection volume 0.1ul, pressure 0.8 psi, total flow-183 Ml/min, septum purge flow-1ml/min, split ratio 150:1, air-200 ml/min, h2-20ml/min, he- 15ml/min.
Washing solvent-ethanol .
Injection solvent- benzene and isooctane standard
-
- Posts: 404
- Joined: Wed Nov 09, 2016 12:43 am
You didn't include the Hydrogen flow going to the FID? It will need this as well as Air/Oxygen to ignite the flame.
-
- Posts: 927
- Joined: Mon Feb 13, 2006 6:28 pm
What value of "Lit off" have you set ?
-
- Posts: 5433
- Joined: Thu Oct 13, 2005 2:29 pm
Your carrier gas flows and pressures are really strange; you only have about 1.2 ml/min going down the column, which is much too little for a 530 um diameter column. It would be about right for a 250 um column. Where did the method come from ? Have you used it before ?
Depending what you are injecting (and 0.1 ul sample with a 150:1 split suggests it is something strange) the very low inlet pressure and carrier flow mean that the pressure pulse as the sample flash vaporizes will cause a flow pulse down the column, which might blow out the flame.
Peter
Depending what you are injecting (and 0.1 ul sample with a 150:1 split suggests it is something strange) the very low inlet pressure and carrier flow mean that the pressure pulse as the sample flash vaporizes will cause a flow pulse down the column, which might blow out the flame.
Peter
Peter Apps
8 posts
Page 1 of 1
Who is online
In total there are 11 users online :: 0 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 11 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 11 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science