- Posts: 3
- Joined: Thu Aug 03, 2017 2:23 pm
API is dissolved in mobile phase
weak and strong wash solution = mobile phase
System:
UPLC Acquity (BSM/SampleManager/ColumnHeaterCooler/PDA)
Column: ACE 3 C18, 150 x 4.6 mm
Detection: 240 nm
Flow: 1.0 ml/min
Temp: 30°C
InjVol: 5 µL
Mobile phase:
A (26 parts):
2 g sodium octane sulfonate are dissolved in 300 mL water, pH is adjusted with acetic acid to 3.5
B (74 parts):
methanol
the API solution is injected four times:
- Inj 1
- 30 other injections
- Inj 2
- 30 other injections
- Inj 3
- 30 other injections
- Inj 4
Areas of the API solution:
Inj 1: 100'000
Inj 2: 96'000
Inj 3: 94'000
Inj 4: 92'000
If the vials, the column, the wash solutions and the mobile phase components are transferred to an other UPLC system and ran on this second system, the pattern is the same.
What effect might cause this phenomenon?
Thank you very much for your help
Florian