Hi folks,

We recently injected a solution of Nicarbazin (in DMF) with a mobile phase of acetonitrile and phosphate buffer (pH 2.8) onto C18 column and saw two peaks...one small followed by one large. PDA scan of both suggested both are same or similar compounds.

Looking at our validation data, I noticed all was fine. We ran the same sample on two other machines with the same outcome. Is the compound sticking to the sampling needle? Even changing the rinse solvent to pure DMF (from 90/10 Water/MeOH) did not help. Bad injector seals perhaps? Or could it be something to do with the column, at the inlet end?

Any ideas? Thanks in advance.

You may be seeing the "strong diluent" effect. When the sample is dissolved in something stronger than the initial mobile phase, the sample tends to get "smeared" into the column instead of equilibrating nicely. The symptom can be tailing, fronting, splitting, shoulders, etc, or even a secondary peak.

If possible, try to dilute your sample with the mobile phase and see if the problem goes away.

If it doesn't, what's the possibility that you actually have two compounds with similar chromophores (UV spectra in solution are notoriously information-poor)?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Hi,

I guess the DILUENT is creating problems for you. The stronger the diluent as compared to the mobile phase the more is the possibility of it not getting equilibrated. Try using the same or rather similar DILUENT (in composition) to the mobile phase. May be the problem will minimise

Chreers,

Yogesh

Thanks Tom and Yogesh, appreciate your excellent suggestions.

I also noticed the two peaks in the blank sample (which is just DMF). Sequential injection of consecutive blank samples saw the intensity of these two peaks gradually diminished. Very strange indeed....perhaps Tom is correct in that there really are 2 closely related compounds/chromophores for Nicarbazin.

Nicarbazin IS a 1:1 of complex of two compounds. You should get two peaks.

Thanks Uwe.....I just found that out. Nicarbazin is actually a 1:1 mixture of 4,4'-dinitrocarbanilide (DNC) and 2-hydroxy-4,6-dimethyl pyrimidine; DNC being the main active drug. Hence, Nicarbazin is usually assayed in relation to DNC.

Still, I need to figure out what's going on in my blank injections.
6 posts Page 1 of 1

Who is online

In total there are 25 users online :: 1 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Amazon [Bot] and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry