Problems with THF? (contd.)
Posted: Wed Oct 11, 2006 12:09 pm
This post is the continuation of two former topics (24 Mar. 2006, "Mobile phase with acetonitrile, tetrahydrofurane and water", and 25 jul "THF quality in HPLC analysis".
Basically the problem is that a newly developed method that uses Inertsil ODS3 column, CH3CN : THF : H2O = 21 : 13 : 66 (v/v/v) mobile phase, 242nm, 40ºC does not work, since the peaks get splitted after a few injections.
The strange thing is that after the development we validated the method without any problem (about 3-4 weeks' day and night work), but after the validation the peak splitting appeared. After an investigation on the solvent batches, it seemed that the whole validation was performed using one single THf batch, and when that THF batch ended, the problem appeared.
At the end of July / beginning of August I received on the Chromforum some ideas to experiment, but at that time had no possibility to do it because of holiday and other projects.
Now I have some new information on the topic that would like to share and would like to get your oppinion / help on the issue.
1. The THf used during the development and validation was from Merck. We now tried THF from Labscane, Lancaster, aldrich and Chromsolv, sooner or later all of them resulting in the same peak splitting phenomena (sometimes after 15-20 injections, sometimes after 5 injections or less).
During these experiments we also noticed that sometimes the peak profile get worst with each injection, then suddenly from one injection to the next one get "normal". This "normality" only maintained for 1-2 injection, getting worst again and at the end splitting the peak.
2. We filtered THF (from Aldrich) through a C18 column and prepared mobile phase with this "filtrated" THF.
The result was that the peaks were splitted right in the first injection. We confirmed this fact with three different columns.
3. We distiiled THF (from Aldrich) and prepared mobile phase with the distilled THF. Sooner or lated the peaks got splitted again. (two tentatives, first one after 6 injections, second after 12-15.)
The first tentative was conducted with the column used for the THF filtration.
4. We set a precolumn between the pump and the injector and tried our luck. Unfortunately the peaks splitted again after 12 injections.
5. Since for all the above mentioned experiments we injected only resolution mixture, another experiment was to perform several blank injections between the resolution injections, to see if the problem came from the sample or from the mobile phase.
We injected once the resolution, three times the blank (CH3CN : H2O) and then the resolution again. The peaks were splitted.
6. As I explained in my previous post, this new method was developed to substitute two other methods that are currenlty used for the analysis of the respective product. One of the current methods is a CH3CN : H2O gradient and the other is a THF : H2O = 25 : 75 isocratic method.
Traditionally the isocratic method have not given problems, but we tried to perform it to see if the problem is the mixture of the three solvents or it is present when the mixture of only two solvents (THf and water) is used.
using Chromsolv THf we were able to perform 10 injections without splitting the peak (we did not perform more injections because of the week end), then the same column using THF from Merck splitted the peaks in 15 injections.
I have to note that all columns used were old columns that previously had been used for other projects, but considering the type of the problem, we obviously did not want to destroy new columns.
These were the experiments that we have made, unfortunately without solving the problem, therefore I am looking forward to receiving your opinion. If anybody is interested, I can send chromatograms or overlays by email.
Thanks in forward,
Kati
P.S. It seems very strange that all over the world only we have this problem, only we are not able to perform a CH3CN / THf / H2O analysis. What is happening here???
Basically the problem is that a newly developed method that uses Inertsil ODS3 column, CH3CN : THF : H2O = 21 : 13 : 66 (v/v/v) mobile phase, 242nm, 40ºC does not work, since the peaks get splitted after a few injections.
The strange thing is that after the development we validated the method without any problem (about 3-4 weeks' day and night work), but after the validation the peak splitting appeared. After an investigation on the solvent batches, it seemed that the whole validation was performed using one single THf batch, and when that THF batch ended, the problem appeared.
At the end of July / beginning of August I received on the Chromforum some ideas to experiment, but at that time had no possibility to do it because of holiday and other projects.
Now I have some new information on the topic that would like to share and would like to get your oppinion / help on the issue.
1. The THf used during the development and validation was from Merck. We now tried THF from Labscane, Lancaster, aldrich and Chromsolv, sooner or later all of them resulting in the same peak splitting phenomena (sometimes after 15-20 injections, sometimes after 5 injections or less).
During these experiments we also noticed that sometimes the peak profile get worst with each injection, then suddenly from one injection to the next one get "normal". This "normality" only maintained for 1-2 injection, getting worst again and at the end splitting the peak.
2. We filtered THF (from Aldrich) through a C18 column and prepared mobile phase with this "filtrated" THF.
The result was that the peaks were splitted right in the first injection. We confirmed this fact with three different columns.
3. We distiiled THF (from Aldrich) and prepared mobile phase with the distilled THF. Sooner or lated the peaks got splitted again. (two tentatives, first one after 6 injections, second after 12-15.)
The first tentative was conducted with the column used for the THF filtration.
4. We set a precolumn between the pump and the injector and tried our luck. Unfortunately the peaks splitted again after 12 injections.
5. Since for all the above mentioned experiments we injected only resolution mixture, another experiment was to perform several blank injections between the resolution injections, to see if the problem came from the sample or from the mobile phase.
We injected once the resolution, three times the blank (CH3CN : H2O) and then the resolution again. The peaks were splitted.
6. As I explained in my previous post, this new method was developed to substitute two other methods that are currenlty used for the analysis of the respective product. One of the current methods is a CH3CN : H2O gradient and the other is a THF : H2O = 25 : 75 isocratic method.
Traditionally the isocratic method have not given problems, but we tried to perform it to see if the problem is the mixture of the three solvents or it is present when the mixture of only two solvents (THf and water) is used.
using Chromsolv THf we were able to perform 10 injections without splitting the peak (we did not perform more injections because of the week end), then the same column using THF from Merck splitted the peaks in 15 injections.
I have to note that all columns used were old columns that previously had been used for other projects, but considering the type of the problem, we obviously did not want to destroy new columns.
These were the experiments that we have made, unfortunately without solving the problem, therefore I am looking forward to receiving your opinion. If anybody is interested, I can send chromatograms or overlays by email.
Thanks in forward,
Kati
P.S. It seems very strange that all over the world only we have this problem, only we are not able to perform a CH3CN / THf / H2O analysis. What is happening here???