I have no ammonium hydroxide so I am running C18 column 150mm x 4mm x 3.5um with 20mM H3PO4 to pH 3.6 with sodium hydroxide. The gradient is 10% to 40% acetonitrile over 10minutes.
Are there any problems associated with using sodium hydroxide in mobile phase?
UV is 210nm.
I seem to get some mixing problems and high baselines sometimes.
WK

May I ask why you are running at pH=3.6? The three pKa of phosphoric acid are 2.1, 7.0 and 12.3.
For straight forward RP separations it should make no difference if the counter ion is sodium or ammonium.


cheers
--
Robert Haefele

In the range of acetonitrile in which you are working there is no solubility problem with either salt. However, you may need to flush the column with water before putting it into a storage solvent.
3 posts Page 1 of 1

Who is online

In total there are 24 users online :: 0 registered, 0 hidden and 24 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 24 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry