Posted: Tue Aug 01, 2017 5:39 pm
by aliyoyo78
Dear all
if anyone could help me. I am doing reaction of methanol to hydrocarbons in a Fixed bed reactor connected online to Shimadzu GC-FID with HP-Plot Q Column. The methanol feed through bubbler with helium (as gas). the problem that I can not detect the peak of methanol . I have heated all line from the reactor to GC to insure that methanol will not condensate. Also I can detect Toluene and Benzes but not methanol. Thank you

Re: Methanol detect bgy FID

Posted: Wed Aug 02, 2017 8:30 pm
by Alfonso
Hello,

My name is Alfonso Madrigal and I work for Advanced Analytical Systems. We specifically provide laboratory support services, such as laboratory instrumentation sales and servicing, monthly maintenance plans including on-site and remote access, troubleshooting and complete instrument repairs, as well as our method development or even our custom laboratory information management systems (LIMS).

AAS specializes in servicing current and older instruments of Varian/Bruker, HP/Agilent, Shimadzu, SRI, and Antek. Our on-site services vary from weekly, monthly and yearly. Additionally, we provide exceptional operator training that improves relations between users and advanced software.

If you have any needs in regard to your laboratory, gas chromatographs, software, or any instrumentation at all, please feel free to contact me with any questions.

Thank you,

Alfonso Madrigal
P: (562) 852-6071
E: amadrigal@aasystems.com

Re: Methanol detect bgy FID

Posted: Sun Aug 06, 2017 10:16 am
by antonk
Seems like methanol is still sitting in your PLOT.
Try to condition you column without sample injection at higher temperatures to determine if the methanol is there or another solutions - use HP/Rtx/Db-5 column suitable both for alcohols and hydrocarbons.