IC columns and mobile phase compatibility
Posted: Tue Oct 10, 2006 12:25 pm
by key4
I´m looking for some information about mobile phase compatibility for IC anion columns, I have found that some silica based IC columns are not compatibles with Methanol or Acetonitrile (Vydac or Dionex for example) while others (alltech) can run up to 100% organic.
Is this due to silica modifications?
regards
K4
Posted: Tue Oct 10, 2006 3:15 pm
by rhaefe
I doubt that the Dionex IC columns are silica based. Solvent compatibility depends on how the ion-exchanger groups are attached to the core particle. These groups can be attached covalently (e.g.Hamilton PRP-X100) or a ion-exchange latex is attached to a sulfonated core particle through electrostatic interaction (most Dionex columns).
It is possible to strip the latex from the substrate. This is not possible with covalently attached ion-exchange groups.
Always refer to the column manual for solvent compatibilty information.
Posted: Fri Oct 13, 2006 12:52 pm
by key4
thanks Robert,
You are rigth Dionex is not based on silica,
thanks for the information on Hamilton PRP
K4