Dear Rod,
First of all, I would like to thank for your interest.
We have met the problem in external std., which has been prepared freshly, using same brand Referance Stds at the same concentrations ( in the past and now).
Areas of TRACE 2000 GC
OVI______________~RT______________~Peak Area
(in ext. Std.)__past____now_______past_______now
MeOH.............7,583......7,610.........4338660......5294288
EtOH..............10,373....10,445.......5811965......10380016
AcON.............12,206....12,298.......20208062.....35394817
MeCl2............14,579....14,712.......3076915.......3420816
n-Hxn............18,109....18,275.......22226917.....43006369
Diisopropyl....19,813....20,022.......84549309.....160953170
ether
Also reproducibility test was performed in an other lab. by using HP GC.
~Peak Area
past______now
85.2..........84,8
154.2........160,6
497.3........471,8
37.6.........50,2
286.7.......491,3
1465,9.....1876,6
In our lab. , there are problems with areas of Ethanol, n-hexan and Diisopropyl ether, but n_Hexan and Diisopropyl ether in the other lab.
When method validation was performed, linearity test was done between 10-150% level of target concentration. It seems; Ethanol, n-hexan and Diisopropyl ether are out of linearity range.
(Note: Repaetability result and reproducibility result are in the acceptance criteria.)