Problem with Aldehyde containing species in LC-MS method
Posted: Fri Oct 06, 2006 8:26 pm
I am trying to develop a method to chromatography 4-carboxybenzoic acid in terephthalic acid. I need the method to be MS friendly.
The chromatogram at:
http://littledomain.com/aldehyde_problem.pdf
shows the bad peak shape I am observing. A gradient separation was used on a Zorbax C18 stablebond column using 0.01 % formic acid/methanol and water. The peak shape looks fine if you increase the concentration of formic acid to 0.1-0.5%.
Unfortunately, the higher concentration of formic acid suppresses the response for lower MW species such as benzoic acid.
The same bad peak shape is noted with 0.5% acetic, 0.01% trifluoroacetic acid, but is not note with 0.14% phosphoric.
Thought it might be due to methanol forming an acetal during separation with the aldehyde group, but same problem is noted with acetonitrile and tetrahydrofuran.
Anyone know why such poor peak shape?
Could it be avoided by switching to a different type C18 reverse phase column?
Derivatization of the aldehyde?
Thanks for the help.