klamtza » Tue May 23, 2017 7:32 am
So first mistake Atropin is of course not a pyrrolizidine alkaloid its a tropane alkaloid.....you be forgiven.
- sorry, of course atropine is a tropane alkaloid
Since i noticed from your picture that you are german i guess you used the published method of the Federal Risk Assessment Institute(BfR) for plant material. The method is not validated for tropane alkaloids so don't use them, sometimes the coeuleted substances can drastically effect the ionization effiency.
- I'm using a modified method based on BfR PA-Tea
See if you can get a matrix matched standard calibration that makes sense, but most likely you need to improve sample preparation, treat it like they treat black tea in the method.
- 10 of 15 TAs show sufficient recoverys (of the same sample work up!) and indicate that sample are prepared and injected correctly
Otherwise is it a possibility for you to use a better separation system? GC could do the trick(there is an excellent method published by Kempf) or a better lc program should deliver reproducable results.
- assuming there is an coeluting interference, how do you explain that the interference influences the spiked and unspiked sample with a different impact?
Also check with an internal standard in your sample and replicates if your lc system might be the problem and doesn't inject it as good as you might think...
- internal standard shows sufficient recovery with good RSDs
- I've injected the sample 6 times in a row to check for possible carry over. RSDs are low
Assuming there is not a problem with your sample prep or putting a vial in the wrong position or something... It is likely due to matrix effects.
- see above, these kind of mistakes can be excluded
There's no easy solution. But I think it would make sense to work in a lower concentration range (dilute everything). Judging by the peak areas you can probably afford to dilute 10-100 times. Matrix effects are less severe in lower concentrations. Another thing you could try is to lower the Q1 and/or Q3 peak width. This makes the detection more selective (narrowing the m/z around the peak), but it also lowers the overall signal.
- I've already diluted samples 1:10, the effect is the same
- matrix effects were compensated
- using a different column also leads to the same efffect
The separation between your compounds of interest might be OK, but there's a forest of peaks that you don't see (you're not measuring them) and they can affect the ionization of your compounds.
- as said above: how can it come that this effects a sample with spike in a different way than a sample without spike?