[quote="EnvironLab"]Just a little question: What happens if your column happens to break inside of the oven near the detector end?[/quote
The main risk is when no column is installed. In the 1980s, during a visit here, Walter Jennings once said that the only explosion he had was when he forgot to hook up a 1/4" glass packed column and turned the pressure-controlled hydrogen carrier on

. The lower explosion limit is 4% by volume, which is difficult to achieve, given that most GC ovens aren't hermetic
Modern GCs have sprung-loaded oven safety vents, or equivalent. On Agilent GCs, the inner door seal is sprung-loaded with securely-latched outer door that will vent when internal pressure rapidly increases, and then reseal when it drops. The danger from a hydrogen explosion is that a closed, sealed oven would disintegrate.
From my experience, even forgetting to put in capillary column for almost an hour at 100 ml/min H2 doesn't seem to cause a bang on a HP5890, or if it did, I didn't notice

.
Bruce Hamilton