Cation exchange chromatography injection solvent issues
Posted: Mon Oct 02, 2006 8:15 pm
Lets say one uses 0.015 N H2SO4 as eluant in mobile phase for cation exchange chromatography. If my rough calculation is correct (and it might not be, correct me if I'm wrong), pH should be around 1.8 to 2. Then suppose one wants to determine Na+ at a ppb level, but the injection solvent has a pH < 1, or potentially 10X or thereabouts more concentrated in hydronium ion than the eluant. Is that in itself a cause for chromatographic problems in finding Na+ at trace levels? It seems that the amount of Na+ in the diluent would be so overwhelmed by the amount of H+ in the diluent, that problems would ensue in detection of Na+. Again, my calculation might be wrong, but I think we'd be dealing with ~ 50 ppm of H+ in an environment where we were looking for Na+ at a ppb level. The Na+ would be "lost in the crowd" of H+ and how would we ever find it? Bottom line - if the pH of the injection media is a good bit lower than that of the eluant, will that cause problems in looking for Na+?