Moving retention time with protein method
Posted: Mon Oct 02, 2006 2:10 pm
I have a RP HPLC method for quantitating a large protein. The separation is done with a Vydac C8, 300A, 5 um, 2.1x250 mm column. The mobile phase is a gradient of ACN/water/0.1% TFA. The gradient starts at 30% ACN and ramps to 92%ACN over 15 min. The main peak comes out at 11.3 min. on my Agilent HP1100 system. When I run this method on my Dionex Ultimate 3000, the peak comes out at 8.6 min. I have to transfer this method to another lab with an older Agilent HP1100 and the retention time on that instrument is about 14 minutes. The general shape of the chromatograms are the same on all instruments.
It was suggested to me that maybe the differences in retention time were caused by differences in instrument dwell time. I added another piece of tubing to my Agilent instrument in front of the column to try to increase the dwell time but I did not see any change relative to the solvent front.
Any ideas of what could cause this peak to come off at such different times on these different instruments?
How big of a deal is it to have different retention times on different instruments if the method is otherwise behaving well (ie. good linearity and precision) on each instrument?
KarenJ
KarenJ