Ion-exchange Chromatography - Fronting and tailing peaks

Posted: Sun Jul 16, 2017 11:33 pm
by ultima_86
Hi everyone, I am new in IC field and I am doing Sodium and Potassium analysis with Shimadzu HPLC system (Conductivity detector) with the following test conditions:

Mobile phase: 0.005M Nitric acid (HPLC grade) diluted with de-ionized water (conductivity less than 0.1 µS). It is filtered through 0.45µm membrane and sonicated for 30 minutes to eliminate all remained bubble.
Column oven: 40 C (there is another component that needs temperature to be kept at 43 C but I do not know what to call it, sorry for my bad English)
Flow rate: 1.0 mL/minute.
My sample are prepared by diluting high-purity Potassium Chloride and Sodium Chloride in the same de-ionized water used for mobile phase preparation.

Here is my problems.
1. Peak due to Sodium is fronting (Asymmetry factor: 0.755), whereas, peak due to Potassium is tailing (Tailing factor: 1.749) although I am using a totally newly-purchased column (Shimadu column IC - C1) which is certified to have great performance with alkaline analysis using test conditions same to mine.

2. Retention time fluctuates day after day although I used the same mobile phase.

Any advice would be highly appreciated. Thanks in advance.

Re: Ion-exchange Chromatography - Fronting and tailing peaks

Posted: Thu Jul 27, 2017 11:32 am
by Khanom
Hi,

What is the detector temperature, this should be the same as the column temperature. A column temperature of 40°C may be too high, try 30 or 35°C.

Has the column been fully conditioned?

What do you mean by ‘alkaline analysis’. You are injecting salts into an acidic mobile phase.

Re: Ion-exchange Chromatography - Fronting and tailing peaks

Posted: Sat Jul 29, 2017 8:14 am
by ultima_86
Hi, Khanom. Thanks for your reply. We didn't make the pre-condition before analysis because pre-condition was not mentioned in column instructions. If possible, can you support me the guideline of column pre-conditioning.

Our sample, which consists of Potassium Sodium hydrogen citrate as active subtance and other excipients, are dissolved in DI-water with the concentration of Sodium+ and Potassium+ is 135ppm and 170 ppm respectively. Due to excessively high concentration which causes peak tailing of Potassium, reducing concentration by 4 times solves tailing problem but fronting still remains. Any suggestion from you are highly appreciated. Thanks in advance!

Re: Ion-exchange Chromatography - Fronting and tailing peaks

Posted: Tue Aug 08, 2017 1:25 pm
by Khanom
Dear Ultima,

Every new column should be conditioned, before it is used for analyses.

As a minimum, pump the mobile phase through the column for 60 minutes at a low flow rate, say 0.5 mL/minute. Many experienced users condition the analytical column at a lower flow rate, overnight for around 12 hours.

Some folk inject a standard onto the column, which often aids in column conditioning.

I am pleased that you have a more Gaussian potassium peak. After you have conditioned the column, if your sodium peak is still asymmetric, please display an image of the whole chromatogram.