Impact of nitrogen generator exhaust on QTOF precision
Posted: Fri Jul 14, 2017 7:29 am
by arnabroy81
My QTOF if showing huge mass deviation, within a run, about 0.05 Da to sometimes 0.5 Da. I learned that QTOF mass measurements depend on temperature stability. The position of our instrument is like this:
https://drive.google.com/file/d/0B4z8B_ ... sp=sharing
The exhaust of the nitrogen generator is pointed towards the back side of QTOF and the rough pump is also at the back. I have noticed the temperature in the back of the QTOF is actually different, little warmer than the rest of the room. My question is, is this a valid potential reason for the observed mass deviation?
However, our service engineer is predicting it might be because of some component problem inside the machine.
Re: Impact of nitrogen generator exhaust on QTOF precision
Posted: Fri Jul 14, 2017 3:51 pm
by James_Ball
Warmer in that area would not be a problem as long as it is fairly consistent temperature. If the temperature is fluctuating several degrees over a few hours then that could be a problem.
Posted: Fri Jul 14, 2017 5:23 pm
by arnabroy81
That means I should put the temperature scale near that area and check. Thanks
Posted: Mon Jul 17, 2017 10:37 am
by lmh
trust the service engineer on this one; they know what happens to their instruments. A deviation as large as 0.5Da can't just be thermal expansion of the flight tube (even 50mDa is pushing the boundaries of believability).